Running GC 8890 / 5977B MSD with Masshunter 10.0 software. We want to analyze samples for ASTM D2887... does Agilent have software to perform this analysis on the chromatograms? Where can it be found / downloaded?
Find maintenance, best practice, and troubleshooting articles.
Running GC 8890 / 5977B MSD with Masshunter 10.0 software. We want to analyze samples for ASTM D2887... does Agilent have software to perform this analysis on the chromatograms? Where can it be found / downloaded?
Hi write2jake,
Currently we do not offer software that can perform SIMDIS analysis on MassHunter mass spec data.
Hi write2jake,
Currently we do not offer software that can perform SIMDIS analysis on MassHunter mass spec data.
Yes, G2887BA is for use with GC ChemStation.
Thanks for your reply.
I was told the part number G2887BA is what I needed by someone in “online chat” through Agilent’s website. Just for clarification, is this part number / software for chem station?
This is great information. Thank you.
So, for the big question: how would one analyze for Simulated Distillation D2887 using MassHunter 10?
There is currently no available solution for working with MassHunter data.
If your GC has an FID in addition to a mass spec, you could use OpenLab GC ChemStation to acquire the FID signal data. Then addon G2887BA could be used to analyze the ChemStation data. You would need to discuss these options with your local sales person.