MassHunter 10.1 Calibration Curve Not Being Created

Hello, 

In the past when creating a calibration curve I would set my cal standards to cal, set the level then go to the method editor and add the cal levels.  Then when pressing Analyze the responses would be automatically added. 

For some reason now the responses are not being auto added and I cannot click into the response area and add the responses manually.  Is there a fix for this or do I have to re-create the batch?

  • Hello  ,

    It is not possible to enter responses manually in the Method Editor. 

    As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels from calibration samples, and then copying those levels to all other samples. If everything else is set up correctly, exiting the method editor and analyzing the batch should create all of the curves. On returning to the method editor the responses should be populated.