Masshunter 12.1 reporting python code help

I am working with the python code file shipped with Masshunter 12.1 entitled "CC_Avg.py". It is referenced by the xml document called "Env_CC_Avg.report.xml".

I am trying to get it to show peak resolution on the report as one of the requirements of EPA 8270E Semivolatiles analysis. The problem is that in the following code example, the cmpd.ResolutionFront attribute does not exist. I believe this is because it is bound to the compound information rather than the peak information.

How can I reference or extract the target peak information so that I can use the ResolutionFront attributes?

I tried to use something like the following without success:

peak = None if compound.IsPrimaryHitPeakIDNull() else DataProvider.GetPeak(istd.BatchID, istd.SampleID, cmpd.CompoundID, cmpd.PrimaryHitPeakID)

ResolutionValue = peak.ResolutionFront

I have also attached the python file itself as a .txt file, starting at line 292.

#CC_Avg.py, Aera% is calcuated against avg resp of the calibrators
#
#import os
import System
import System.IO

import clr
clr.AddReference("iTextSharp")

import iTextSharp
import iTextSharp.text
import iTextSharp.text.pdf

from iTextSharp.text import *
from iTextSharp.text.pdf import *

#-----------------------------------------------------------------------------
# PageEvent
#-----------------------------------------------------------------------------
class PageEvent (PageEventBase):
    def __init__(self,font):
        PageEventBase.__init__(self,font)

    def GetTitle(self, writer):
        return resmgr.GetString("Title")

    def GetFooterCenterText(self, writer):
        format = resmgr.GetString("PageOf")
        return System.String.Format(format, writer.PageNumber)

    def GetFooterRightText(self, writer):
        format = resmgr.GetString("GeneratedAtOn")
        return System.String.Format(format, self.now.ToString("t"), self.now.ToString("d"))

#-----------------------------------------------------------------------------
# mid levels 
#-----------------------------------------------------------------------------
def Mid():
    ccExpC,avg,mid= ({} for i in range(3))
    samples = DataProvider.SelectSamples("","AcqDateTime")   
    if 0 < len(samples):
        sample = samples[-1]
        if not sample.IsCalibrationReferenceSampleIDNull():
            sample = DataProvider.GetSample(sample.BatchID, sample.CalibrationReferenceSampleID)
            targets = DataProvider.SelectCompounds(
                    System.String.Format(
                    System.Globalization.CultureInfo.InvariantCulture,
                    "BatchID={0} AND SampleID={1} AND Not ISTDFlag=True",
                    sample.BatchID, sample.SampleID),
                    "RetentionTime")
                
            CCsample = DataProvider.SelectSamples("SampleType='CC'","AcqDateTime")[0]
            for compound in DataProvider.SelectCompounds(System.String.Format(
                    System.Globalization.CultureInfo.InvariantCulture,
                    "BatchID={0} AND SampleID={1} AND Not ISTDFlag=True",
                    CCsample.BatchID, CCsample.SampleID),
                    "RetentionTime"):
                ccExpC[compound.CompoundName] = 0 if compound.IsExpectedConcentrationNull() else compound.ExpectedConcentration
            for t in targets:
                levelName, diffC, levelRes, ilevelRes = ([] for i in range(4))
                ccExp = ccExpC.get(t.CompoundName, 0)
                istd = None if t.IsISTDCompoundIDNull() else DataProvider.GetCompound(sample.BatchID, sample.SampleID, t.ISTDCompoundID)
                for calibration in DataProvider.SelectCalibrations(
                                System.String.Format(
				System.Globalization.CultureInfo.InvariantCulture,
				"BatchID={0} AND SampleID={1} AND CompoundID={2} AND CalibrationType='Calibration'",
				t.BatchID,
                                t.SampleID,
                                t.CompoundID),
                                "LevelConcentration"):
                    levelName += calibration.LevelName,
                    levelRes += 0 if calibration.IsLevelResponseNull() else calibration.LevelResponse,
                    diffC += abs(calibration.LevelConcentration - ccExp),
                n = diffC.index(min(diffC))
                lName = levelName[n]
                avg[t.CompoundName] = sum(levelRes)/len(levelRes)
                mid[t.CompoundName] = levelRes[n]
                if istd is not None:
                    for cal in DataProvider.SelectCalibrations(
                                System.String.Format(
				System.Globalization.CultureInfo.InvariantCulture,
				"BatchID={0} AND SampleID={1} AND CompoundID={2} AND CalibrationType='Calibration'",
				istd.BatchID,
                                istd.SampleID,
                                istd.CompoundID),
                                ""):
                        ilevelRes += 0 if cal.IsLevelResponseNull() else cal.LevelResponse,
                        if cal.LevelName==""+lName+"":
                            mid[istd.CompoundName] = 0 if cal.IsLevelResponseNull() else cal.LevelResponse
                    avg[istd.CompoundName] = 0 if len(ilevelRes)==0 else sum(ilevelRes)/len(ilevelRes)
                    
    return avg, mid, ccExpC                        
#-----------------------------------------------------------------------------
# count cal levels
#-----------------------------------------------------------------------------
def CountCalLevels(compound):
    pathName,acqTime,levelName,CCFiles,CCAcqTime,CCLevel = ([] for i in range(6))

    for calibration in DataProvider.SelectCalibrations(
                                System.String.Format(
				System.Globalization.CultureInfo.InvariantCulture,
				"BatchID={0} AND SampleID={1} AND CompoundID={2} AND Not CalibrationType='Method'",
				compound.BatchID,
				compound.SampleID,
                                compound.CompoundID),
                                "LevelConcentration"):
        if calibration.CalibrationType == "CC":
            CCFiles += "" if calibration.IsCalibrationSTDPathNameNull() else calibration.CalibrationSTDPathName,
            CCAcqTime += "" if calibration.IsCalibrationSTDAcquisitionDateTimeNull() else calibration.CalibrationSTDAcquisitionDateTime,
            CCLevel += "" if calibration.IsLevelNameNull() else calibration.LevelName,
        if calibration.CalibrationType == "Calibration":
            pathName += "" if calibration.IsCalibrationSTDPathNameNull() else calibration.CalibrationSTDPathName,
            acqTime += "" if calibration.IsCalibrationSTDAcquisitionDateTimeNull() else calibration.CalibrationSTDAcquisitionDateTime,
            levelName += "" if calibration.IsLevelNameNull() else calibration.LevelName,
    return levelName,pathName,acqTime,CCFiles,CCAcqTime,CCLevel
#-----------------------------------------------------------------------------
# count max cal levels
#-----------------------------------------------------------------------------
def CountMaxCalLevels(sample):
    pathName,acqTime,levelName,CCFiles,CCAcqTime,CCLevel,levelList = ([] for i in range(7))   

    compounds = DataProvider.SelectCompounds(
            System.String.Format(
            System.Globalization.CultureInfo.InvariantCulture,
            "BatchID={0} AND SampleID={1}",
            sample.BatchID, sample.SampleID),
            "RetentionTime")
    for compound in compounds:
        levelName,pathName,acqTime,CCFiles,CCAcqTime,CCLevel = CountCalLevels(compound)
        levelList += len(levelName),
    cmpd = compounds[levelList.index(max(levelList))]    
    levelName,pathName,acqTime,CCFiles,CCAcqTime,CCLevel = CountCalLevels(cmpd)       
    return levelName,pathName,acqTime,CCFiles,CCAcqTime,CCLevel
#-----------------------------------------------------------------------------
# cal Info
#-----------------------------------------------------------------------------
def BatchInfo(sample):
    arr = System.Array[System.Single]([1.1,6])
    table = PdfPTable(arr)
    table.WidthPercentage = 100
    table.DefaultCell.Border = 0
    table.HorizontalAlignment = Element.ALIGN_LEFT
    table.SpacingAfter = 8            
    table.AddCell(Phrase(resmgr.GetString("BatchName"), fontBold))
    table.AddCell(Phrase(DataProvider.BatchAttributes.BatchDataPathFileName, font))
    table.AddCell(Phrase(resmgr.GetString("MethodFile"), fontBold))
    table.AddCell(Phrase("" if sample.IsDAMethodFileNameNull() else sample.DAMethodPathName + "\\"+ sample.DAMethodFileName, font))
    table.AddCell(Phrase(resmgr.GetString("DailyCC"), fontBold))
    table.AddCell(Phrase(sample.DataPathName.ToString()+"\\"+sample.DataFileName.ToString(), font))
    document.Add(table)
  
def ProcessSamples():
    samples = DataProvider.SelectSamples("","AcqDateTime")   
    if 0 < len(samples):
        sample = samples[-1]
        if not sample.IsCalibrationReferenceSampleIDNull():
            sample = DataProvider.GetSample(sample.BatchID, sample.CalibrationReferenceSampleID)
            levelName,pathName,acqTime,CCFiles,CCAcqTime,CCLevel = CountMaxCalLevels(sample)
            CCsamples = DataProvider.SelectSamples("SampleType='CC'","AcqDateTime")
            if len(CCsamples)==0 or len(CCFiles) ==0:
                document.Add(Paragraph(resmgr.GetString("NoCC2"),fontBold))
            else:
                CCsample = CCsamples[0]
                BatchInfo(CCsample)
## cal files section:
                arr = System.Array[System.Single]([1,1.3,4.7])
                table1 = PdfPTable(arr)
                table1.WidthPercentage = 100
                table1.DefaultCell.Border = 0
                table1.HorizontalAlignment = Element.ALIGN_LEFT
                table1.SpacingAfter = 10
            
                table1.AddCell(Phrase(resmgr.GetString("LevelName"), fontBold))
                table1.AddCell(Phrase(resmgr.GetString("InjectionTime"), fontBold))
                table1.AddCell(Phrase(resmgr.GetString("CalFile"), fontBold))             

                for i in range(len(levelName)):    
                    table1.AddCell(Phrase(levelName[i].ToString(), font))
                    table1.AddCell(Phrase(acqTime[i].ToString(), font))
                    table1.AddCell(Phrase(pathName[i], font))
                if len(CCFiles)>0:
                    table1.AddCell(Phrase(CCLevel[0].ToString(), font))
                    table1.AddCell(Phrase(CCAcqTime[0].ToString(), font))
                    #table1.AddCell(Phrase(CCFiles[0]+"    <======", font)) 
                    table1.AddCell(Phrase(CCFiles[0], font))
                document.Add(table1)

                avg, mid, ccExpC = Mid()      
                targets = DataProvider.SelectCompounds(
                    System.String.Format(
                    System.Globalization.CultureInfo.InvariantCulture,
                    "BatchID={0} AND SampleID={1} AND Not ISTDFlag=True",
                    sample.BatchID, sample.SampleID),
                    "RetentionTime")           

                istdIDs, istds, ntargets, t = ([] for i in range(4))             
                for c in targets:
                    if c.IsISTDCompoundIDNull():
                        ntargets += c,
                    else:    
                        istdIDs += c.ISTDCompoundID,
                        t += c,
                if istdIDs:
                    istdIDs = sorted(set(istdIDs))
                    for i in istdIDs:
                        istd = DataProvider.GetCompound(sample.BatchID, sample.SampleID, i)
                        istds += istd,
            
                CCComps = DataProvider.GetAllCompounds(CCsample.BatchID, CCsample.SampleID)
                CCRespD,CCConcD,CCExpConcD,CCAccuracyD,CCAvgRF = ({} for i in range(5))
                for CC in CCComps:
                    CCPeak = None if CC.IsPrimaryHitPeakIDNull() else DataProvider.GetPeak(
                            CC.BatchID, CC.SampleID, CC.CompoundID, CC.PrimaryHitPeakID)
                    CCRespD[CC.CompoundName] = 0 if CCPeak is None or CCPeak.IsTargetResponseNull() else CCPeak.TargetResponse
                    CCConcD[CC.CompoundName] = 0 if CCPeak is None or CCPeak.IsFinalConcentrationNull() else CCPeak.FinalConcentration
                    CCAccuracyD[CC.CompoundName] = 0 if CCPeak is None or CCPeak.IsAccuracyNull() else CCPeak.Accuracy
                    CCExpConcD[CC.CompoundName] = 0 if CC.IsExpectedConcentrationNull() else CC.ExpectedConcentration
                    CCAvgRF[CC.CompoundName] = 0 if CC.IsAverageResponseFactorNull() else CC.AverageResponseFactor
                    
## istd section:
                if istds:
                    arr = System.Array[System.Single]([3,1,1,1,1,1])
                    table2 = PdfPTable(arr)
                    table2.WidthPercentage = 100
                    table2.DefaultCell.Border = 0
                    table2.HorizontalAlignment = Element.ALIGN_LEFT
                    table2.SpacingAfter = 10
                    table2.AddCell(Phrase(resmgr.GetString("ISTDCompound"), fontBold))
                    table2.AddCell(Phrase(resmgr.GetString("AvgResp"), fontBold))
                    table2.AddCell(Phrase(resmgr.GetString("MidResp"), fontBold))
                    table2.AddCell(Phrase(resmgr.GetString("CCResp"), fontBold))
                    table2.AddCell(Phrase(resmgr.GetString("Area%"), fontBold))
                    table2.AddCell(Phrase("A/M", fontBold))

                    for istd in istds:
                        iavg = avg.get(istd.CompoundName, 0)
                        imid = mid.get(istd.CompoundName, 0)
                        area_A = 0 if iavg==0 else 100*CCRespD[istd.CompoundName]/imid
                        table2.AddCell(Phrase("" if istd is None or istd.IsCompoundNameNull() else istd.CompoundName, font))
                        table2.AddCell(Phrase("" if iavg==0 else iavg.ToString("F0"),font))
                        table2.AddCell(Phrase(imid.ToString("F0"), font))
                        table2.AddCell(Phrase(CCRespD[istd.CompoundName].ToString("F0"), font))
                        if not istd.IsISTDResponseMinimumPercentDeviationNull() and CCRespD[istd.CompoundName] < imid*(istd.ISTDResponseMinimumPercentDeviation+100)/100:
                            table2.AddCell(Phrase("" if area_A == 0 else area_A.ToString("F2")+" #", fontRed))
                        elif not istd.IsISTDResponseMaximumPercentDeviationNull() and CCRespD[istd.CompoundName] > imid*(istd.ISTDResponseMaximumPercentDeviation+100)/100:
                            table2.AddCell(Phrase("" if area_A == 0 else area_A.ToString("F2")+" #", fontRed))    
                        else:
                            table2.AddCell(Phrase("" if area_A == 0 else area_A.ToString("F2"), font))
                        table2.AddCell(Phrase("M", font))
                    document.Add(table2)
            
# Target section: 
                    arr = System.Array[System.Single]([2.5,0.9,0.9,1,0.9,0.9,0.9,1,0.8,1.2])
                    table3 = PdfPTable(arr)
                    table3.WidthPercentage = 100
                    table3.DefaultCell.Border = 0
                    table3.HorizontalAlignment = Element.ALIGN_LEFT
                    table3.SpacingAfter = 10
                    table3.HeaderRows = 1
                    table3.AddCell(Phrase(resmgr.GetString("TargetCompound"), fontBold))
                    table3.AddCell(Phrase("Type", fontBold))
                    table3.AddCell(Phrase("Resolution", fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("AvgRF/R2"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CCRF"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("ExpConc"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CalcConc"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("%Dev"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("Area%"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CurveFit"), fontBold))

                    for istd in istds:
                        table3.AddCell(Phrase("" if istd is None or istd.IsCompoundNameNull() else istd.CompoundName, font))
                        cell = PdfPCell(Phrase("------------------------------------ISTD----------------------------------------",font))
                        cell.Colspan = 9
                        cell.Border = 0
                        table3.AddCell(cell)
                        for cmpd in t:
                            if istd.CompoundID == cmpd.ISTDCompoundID:
                                rf = CCRespD[cmpd.CompoundName]*cmpd.ISTDConcentration/(CCRespD[istd.CompoundName]*CCExpConcD[cmpd.CompoundName])                        
                                rfDev = 0 if CCAvgRF[cmpd.CompoundName] is None else 100*(CCAvgRF[cmpd.CompoundName]-rf)/CCAvgRF[cmpd.CompoundName]
                                table3.AddCell(Phrase("" if cmpd.IsCompoundNameNull() else cmpd.CompoundName,font))
                                #I added this code to add compound types for 8270C. Working fine
                                try:
                                    if "CCC" in cmpd.CompoundGroup:
                                        table3.AddCell(Phrase("CCC", font))
                                    elif "Performance Check" in cmpd.CompoundGroup:
                                        table3.AddCell(Phrase("P", font))
                                    else:
                                        table3.AddCell(Phrase(""))
                                except:
                                    table3.AddCell(Phrase(""))
                                #
                                #
                                #
                                #
                                ##Code that runs but doesn't give me the information I need, just placeholder values
                                
                                ResolutionValue = "-"
                                try:
                                    if cmpd.ResolutionLimit is not None:
                                        ResolutionValue = "ResolutionFront"
                                except:
                                    pass
                                try:
                                    if cmpd.ResolutionLimit is not None:
                                        ResolutionValue = "ResolutionRear"
                                except:
                                    pass
                                table3.AddCell(Phrase(ResolutionValue, font))

                                ##Code I would like to work if cmpd had the ResolutionFront and Rear attributes
                                
                                #ResolutionValue = "-"
                                #try:
                                #    if cmpd.ResolutionFront is not None and cmpd.ResolutionFront > 0:
                                #        ResolutionValue = cmpd.ResolutionFront
                                #except:
                                #    pass
                                #try:
                                #    if cmpd.ResolutionRear is not None and cmpd.ResolutionRear > 0:
                                #        ResolutionValue = cmpd.ResolutionRear
                                #except:
                                #    pass
                                #table3.AddCell(Phrase(ResolutionValue, font))                                
                                #
                                #
                                #
                                #
                                    
                                if cmpd.CurveFit == 'fitAverageOfResponseFactors':
                                    table3.AddCell(Phrase("" if CCAvgRF[cmpd.CompoundName] is None else CCAvgRF[cmpd.CompoundName].ToString("F3"), font))
                                    if not CC.IsMinimumCCRelativeResponseFactorNull() and CC.MinimumCCRelativeResponseFactor > rf:
                                        table3.AddCell(Phrase(rf.ToString("F3")+" #", fontBlue))
                                    else:     
                                        table3.AddCell(Phrase(rf.ToString("F3"), font))
                                    table3.AddCell(Phrase("" if CCExpConcD[cmpd.CompoundName]==0 else CCExpConcD[cmpd.CompoundName].ToString("F2"), font))
                                    if cmpd.IsCalibrationReferenceCompoundIDNull():
                                        table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                    else:
                                        table3.DefaultCell.BackgroundColor = BaseColor(221,221,221)
                                        table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                        table3.DefaultCell.BackgroundColor = BaseColor(255,255,255)                                             
                                    if not cmpd.IsMaximumCCResponseFactorDeviationNull() and CCAvgRF[cmpd.CompoundName] is not None and cmpd.MaximumCCResponseFactorDeviation< abs(rfDev):        
                                        table3.AddCell(Phrase("" if rfDev == 0 else rfDev.ToString("F2")+" #",fontRed))                                                                                                                           
                                    else:
                                        table3.AddCell(Phrase("" if rfDev == 0 else rfDev.ToString("F2"),font))

                                else:
                                    table3.AddCell(Phrase("" if cmpd.IsCurveFitR2Null() else cmpd.CurveFitR2.ToString("F4"), fontBlue))
                                    table3.AddCell(Phrase(rf.ToString("F4") if rf<100 else rf.ToString("F0"), font))
                                    table3.AddCell(Phrase("" if CCExpConcD[cmpd.CompoundName]==0 else CCExpConcD[cmpd.CompoundName].ToString("F2"), font))
                                    if cmpd.IsCalibrationReferenceCompoundIDNull():
                                        table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                    else:
                                        table3.DefaultCell.BackgroundColor = BaseColor(221,221,221)
                                        table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                        table3.DefaultCell.BackgroundColor = BaseColor(255,255,255)                                        
                                    AcuracyDev = 100-CCAccuracyD[cmpd.CompoundName]
                                    if not cmpd.IsAccuracyMaximumPercentDeviationNull() and cmpd.AccuracyMaximumPercentDeviation< abs(AcuracyDev):                   
                                        table3.AddCell(Phrase(AcuracyDev.ToString("F2")+" #", fontRed))                                                                 
                                    else:
                                        table3.AddCell(Phrase(AcuracyDev.ToString("F2"), font))                                      
				if (100*CCRespD[cmpd.CompoundName]/mid[cmpd.CompoundName]) > 50 and (100*CCRespD[cmpd.CompoundName]/mid[cmpd.CompoundName]) < 200:
					table3.AddCell(Phrase((100*CCRespD[cmpd.CompoundName]/mid[cmpd.CompoundName]).ToString("F2"),font))
				else:
					table3.AddCell(Phrase((100*CCRespD[cmpd.CompoundName]/mid[cmpd.CompoundName]).ToString("F2"),fontRed))
                              
                                cf = DataProvider.TranslateEnumValue("TargetCompound","CurveFit",cmpd["CurveFit"])
                                #table3.AddCell(Phrase("" if cmpd.IsCurveFitNull() else "Avg RF" if cf=="Average of Response Factors" else cf,font))
				if cmpd.IsCurveFitNull():
					table3.AddCell(Phrase("",font))
				elif cf=="Average of Response Factors":
					table3.AddCell(Phrase("Avg RF",font))
				else:
					table3.AddCell(Phrase(cf,fontBlue))
                    document.Add(table3)
                    document.Add(Paragraph(resmgr.GetString("Note"), fontBold))
##for non-ISTD batch:
                if ntargets:    
                    arr = System.Array[System.Single]([3,1,0.9,0.9,0.9,1,0.8,1.2])
                    table3 = PdfPTable(arr)
                    table3.WidthPercentage = 100
                    table3.DefaultCell.Border = 0
                    table3.HorizontalAlignment = Element.ALIGN_LEFT
                    table3.SpacingAfter = 10
                    table3.SpacingBefore = 10
                    table3.HeaderRows = 1
                    table3.AddCell(Phrase(resmgr.GetString("TargetCompound"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("AvgRF/R2"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CCRF"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("ExpConc"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CalcConc"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("%Dev"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("Area%"), fontBold))
                    table3.AddCell(Phrase(resmgr.GetString("CurveFit"), fontBold))

                    for cmpd in ntargets:
                        rf = CCRespD[cmpd.CompoundName]/CCExpConcD[cmpd.CompoundName]                       
                        rfDev = 0 if CCAvgRF[cmpd.CompoundName] is None else 100*(CCAvgRF[cmpd.CompoundName]-rf)/CCAvgRF[cmpd.CompoundName]
                        table3.AddCell(Phrase("" if cmpd.IsCompoundNameNull() else cmpd.CompoundName,font))
                        if cmpd.CurveFit == 'fitAverageOfResponseFactors':
                            table3.AddCell(Phrase("" if CCAvgRF[cmpd.CompoundName] is None else CCAvgRF[cmpd.CompoundName].ToString("F4"), font))
                            if not CC.IsMinimumCCRelativeResponseFactorNull() and CC.MinimumCCRelativeResponseFactor > rf:
                                table3.AddCell(Phrase((rf.ToString("F4") if rf<100 else rf.ToString("F0"))+" #", fontBlue))
                            else:     
                                table3.AddCell(Phrase(rf.ToString("F4") if rf<100 else rf.ToString("F0"), font))
                            table3.AddCell(Phrase("" if CCExpConcD[cmpd.CompoundName]==0 else CCExpConcD[cmpd.CompoundName].ToString("F2"), font))
                            if cmpd.IsCalibrationReferenceCompoundIDNull():
                                table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                            else:
                                table3.DefaultCell.BackgroundColor = BaseColor(221,221,221)
                                table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                table3.DefaultCell.BackgroundColor = BaseColor(255,255,255)                                     
                            if not cmpd.IsMaximumCCResponseFactorDeviationNull() and CCAvgRF[cmpd.CompoundName] is not None and cmpd.MaximumCCResponseFactorDeviation< abs(rfDev):        
                                table3.AddCell(Phrase("" if rfDev == 0 else rfDev.ToString("F2")+" #",fontRed))                                                                                                                           
                            else:
                                table3.AddCell(Phrase("" if rfDev == 0 else rfDev.ToString("F2"),font))                                
                        else:
                            table3.AddCell(Phrase("" if cmpd.IsCurveFitR2Null() else cmpd.CurveFitR2.ToString("F4"), font))
                            table3.AddCell(Phrase(rf.ToString("F4") if rf<100 else rf.ToString("F0"), font))
                            table3.AddCell(Phrase("" if CCExpConcD[cmpd.CompoundName]==0 else CCExpConcD[cmpd.CompoundName].ToString("F2"), font))
                            if cmpd.IsCalibrationReferenceCompoundIDNull():
                                table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                            else:
                                table3.DefaultCell.BackgroundColor = BaseColor(221,221,221)
                                table3.AddCell(Phrase("" if CCConcD[cmpd.CompoundName]==0 else CCConcD[cmpd.CompoundName].ToString("F2"),font))
                                table3.DefaultCell.BackgroundColor = BaseColor(255,255,255)                                                          
                            AcuracyDev = 100-CCAccuracyD[cmpd.CompoundName]
                            if not cmpd.IsAccuracyMaximumPercentDeviationNull() and cmpd.AccuracyMaximumPercentDeviation< abs(AcuracyDev):                   
                                table3.AddCell(Phrase(AcuracyDev.ToString("F2")+" #", fontRed))                                                                 
                            else:
                                table3.AddCell(Phrase(AcuracyDev.ToString("F2"), font))

                        table3.AddCell(Phrase((100*CCRespD[cmpd.CompoundName]/avg[cmpd.CompoundName]).ToString("F2"),font))          
                        cf = DataProvider.TranslateEnumValue("TargetCompound","CurveFit",cmpd["CurveFit"])
                        table3.AddCell(Phrase("" if cmpd.IsCurveFitNull() else "Avg RF" if cf=="Average of Response Factors" else cf,font))
                    document.Add(table3)
                    document.Add(Paragraph(resmgr.GetString("Note"), fontBold))
                     
#-----------------------------------------------------------------------------
# AddCell
#-----------------------------------------------------------------------------
def AddCell(table, phrase, align):
	cell = PdfPCell(phrase)
	cell.HorizontalAlignment = align
	cell.Border = table.DefaultCell.Border
	table.AddCell(cell)

#-----------------------------------------------------------------------------
# Font
#-----------------------------------------------------------------------------
fontname = GetDefaultFontNameByCulture(System.Globalization.CultureInfo.CurrentCulture)
RegisterFont(fontname)
defaultFont = FontFactory.GetFont(fontname, BaseFont.IDENTITY_H)
fontBold = Font(defaultFont.BaseFont, 9, Font.BOLD)
font = Font(defaultFont.BaseFont, 9)
fontRed = Font(defaultFont.BaseFont, 9, Font.NORMAL, BaseColor.RED)
fontBlue = Font(defaultFont.BaseFont, 9, Font.NORMAL, BaseColor.BLUE)

#----------------------------------------------------------------------------
# Resource manager
#-----------------------------------------------------------------------------
# load resource assembly
resasm = System.Reflection.Assembly.LoadFrom(ResourceAssemblyPath)
# create resource manager
resmgr = System.Resources.ResourceManager("CCal", resasm)

#-----------------------------------------------------------------------------
# Main
#-----------------------------------------------------------------------------
DataProvider.CheckAbortSignal()
stream = System.IO.FileStream(OutputFilePath, System.IO.FileMode.Create, System.IO.FileAccess.ReadWrite)
document = Document()
writer = PdfWriter.GetInstance(document, stream)
pageEvent = PageEvent(defaultFont)
writer.PageEvent = pageEvent
document.SetPageSize(PreferredPageSize)
document.SetMargins(25,25,30,30)
document.Open()

ProcessSamples()
pageEvent.ProcessTotalNumberOfPages(writer)

document.Dispose()
writer.Dispose()

Parents
  • I was able to figure it out on my own. Here is the code for anyone in the future who needs to solve this same issue:

    peak = None if cmpd.IsPrimaryHitPeakIDNull() else DataProvider.GetPeak(istd.BatchID, istd.SampleID, cmpd.CompoundID, cmpd.PrimaryHitPeakID)

    try:
    if peak.ResolutionFront>0:
    if peak.ResolutionFront>50:
    table3.AddCell(Phrase(peak.ResolutionFront.ToString("F0")+"%",fontRed))
    else:
    table3.AddCell(Phrase(peak.ResolutionFront.ToString("F0")+"%",font))
    elif peak.ResolutionRear>0:
    if peak.ResolutionRear>50:
    table3.AddCell(Phrase(peak.ResolutionRear.ToString("F0")+"%",fontRed))
    else:
    table3.AddCell(Phrase(peak.ResolutionRear.ToString("F0")+"%",font))
    else:
    table3.AddCell(Phrase("-",font))
    except:
    table3.AddCell(Phrase("-",font))

Reply
  • I was able to figure it out on my own. Here is the code for anyone in the future who needs to solve this same issue:

    peak = None if cmpd.IsPrimaryHitPeakIDNull() else DataProvider.GetPeak(istd.BatchID, istd.SampleID, cmpd.CompoundID, cmpd.PrimaryHitPeakID)

    try:
    if peak.ResolutionFront>0:
    if peak.ResolutionFront>50:
    table3.AddCell(Phrase(peak.ResolutionFront.ToString("F0")+"%",fontRed))
    else:
    table3.AddCell(Phrase(peak.ResolutionFront.ToString("F0")+"%",font))
    elif peak.ResolutionRear>0:
    if peak.ResolutionRear>50:
    table3.AddCell(Phrase(peak.ResolutionRear.ToString("F0")+"%",fontRed))
    else:
    table3.AddCell(Phrase(peak.ResolutionRear.ToString("F0")+"%",font))
    else:
    table3.AddCell(Phrase("-",font))
    except:
    table3.AddCell(Phrase("-",font))