Index-13 is out of bounds

We get the error "Index-13 is out of bounds" after each sample in the sequence and just doesn't go to another sample. We restarted both the computer and the instrument million times already. Tried to change the path location/method, decreased/increased amount of samples in the sequence (so it happens even when we have just 1 sample in the sequence), created new sequence... Still the same. Please suggest what we can do, as we have to restart the software after each sample run, it's so annoying... Thank you in advance

Parents
  • Index xxx is out of bounds is typically related to the print area/page size requirement in your printer.  I would first try selecting a printer that can accommodate the print requirements, A4 or Letter.  If that doesn't work, you could try to Print to PDF.  If that doesn't work, change your method to acquire the data and not process it and the process it after the sequence has completed.

    And please let us know what happens.

Reply
  • Index xxx is out of bounds is typically related to the print area/page size requirement in your printer.  I would first try selecting a printer that can accommodate the print requirements, A4 or Letter.  If that doesn't work, you could try to Print to PDF.  If that doesn't work, change your method to acquire the data and not process it and the process it after the sequence has completed.

    And please let us know what happens.