Good afternoon,
Good afternoon,
Hello Elmarie,
I apologize for the delayed response. Unfortunately this can be an issue with Optimizer. One possible cause can be having a space somewhere in the data file path name. While this is allowed by the OS and MassHunter in general, in some cases it can cause issues with Optimizer report generation. So avoid using 'D:\MassHunter\Data\My Files\Optimizer' and use 'D:\MassHunter\Data\MyFiles\Optimizer'.
If this is not the issue then one workaround is to use the attached Windows batch file, which for download purposes is attached as a text (.txt) file. Download this and change the extension to .bat so it can be run as a Windows batch file. When this batch file is run it will ask for the Report folder path. You can copy this directly from the Windows file explorer by going to the folder that contains the Reports folder and then while holding the shift key right click on the Reports folder. You will have an option to Copy as path. Once you have copied the path, right click in the batch command window, choose Paste, and press Enter. If you now go back to the report folder the .pdf file should be present.
set /p reportdirectory=Paste the \report directory path here: START C:\"Program Files (x86)"\Agilent\MassHunter\Workstation\Acq\bin\AgtQQQOptimizerReport.exe "%reportdirectory%"