Masshunter - Can you use another compound for calibration if you don't have a calibrator for the specific compound (As you would use another compounds internal standard for quantification)

Good day,

I have set up a new method with about 40 compounds. I have also setup  the quantification method in Masshunter Quant.

I have calibrators for all my compounds except for three of the compounds.

Is it possible to use one of the other compound's calibrators for these compounds - I have gone through all the settings in masshunter- but cannot find I place where such a setting can be made.

Best Regards,

Tania

Parents
  • Hi Tania,

     

    It is possible to apply the calibration curve from one compound to another. There are a couple ways of us to do it depending on your version.

     

    The first way is to copy calibration levels from one compound to another. To do this, go into the Method Editor - Calibration Curve Setup. Left-click on a compound with a calibration curve to select the compound and then right-click to choose Copy Calibration Levels To. A window will open where you can select the compound(s) you would like to apply the calibration curve to. This approach cannot be used to copy ISTD, but the relative ISTD can be used instead. It's necessary to Quantitate, instead of Analyze, when presented the option.

     

    If you have a newer version of MassHunter Quant, there is now an Advanced Method Task that makes this even easier. In the Method Editor -> Advanced Tasks -> Inherit Calibration Reference Setup, it's possible to select the compound from which to inherit the calibration curve in a drop down menu. You would then want to validate the method and save it. 

  • Britt,

    Can you give more details (Step by step) on the process using the newer version? I have version 12.0. 

Reply Children
No Data
Was this helpful?