Hi everyone. Can we convert the data file .D from Mass Hunter software to .cdf file.
Thanks a lot for the help.
Find maintenance, best practice, and troubleshooting articles.
Hi everyone. Can we convert the data file .D from Mass Hunter software to .cdf file.
Thanks a lot for the help.
Thank you for the feedback!
Hello Sara,
As you described, one of the tasks that the GCMS Translator software can perform is to convert *.D files that were collected in MSD ChemStation software to a format compatible with the MassHunter Qualitative and Quantitative Data Analysis products.
If you follow the procedure outlined by Tom in his reply, it unlocks a hidden and not fully supported feature that can export the raw data from single quad GCMS *.D files to AIA format. In order to do this, you will just need make a copy of the GCMS translator desktop shortcut and then edit the properties of the shortcut (right-click>Properties) as described in Tom's instructions and shown below:
Once you apply this edit to the shortcut properties, the next time you launch the GCMS Translator with that shortcut the software will allow you to export raw data to AIA format.
Thank you very much! I had successfully translated the *.D files to *.cdf files. Very happy!