Hi,
After install, I noticed a bottle named only as Production # 1255744 Container # 09630. I need to what is in this solution so I can either store it properly or dispose of it.
Best
Hi,
After install, I noticed a bottle named only as Production # 1255744 Container # 09630. I need to what is in this solution so I can either store it properly or dispose of it.
Best
Hi SamT,
this number doesn't exist in our part number ranges. Could you please add a photo of the label/bottle? TIA
Best regards,
Norbert
Hi,
Here are the images.
Hi,
Here are the images.
Thank you, but it's still nothing we sell separately. It may be part of the LC/MS system kit. My LC/MS colleagues might know more.
Sorry
The label is for the bottle and from the bottle manufacturer. It has no relation to the solution in the bottle or any Agilent part number. The small handwriting on the bottle shows 1:500. This most likely indicates it is a dilution of a standard, probably the installation checkout solution. this would be a solution that is largely acetonitrile and some aqueous composition. Dispose of this solution with your hazardous solvent waste.