Hello,
For the 6495C QQQ, are Pump 1 current and Pump 2 current associated with the turbo pump and/or roughing pump? Are there typical values for these currents?
Thank you
Hello,
For the 6495C QQQ, are Pump 1 current and Pump 2 current associated with the turbo pump and/or roughing pump? Are there typical values for these currents?
Thank you
Hi,
p. altiero is correct. The wattage values should go down as the pump spins up to full speed. Ballpark values when pump is at 100%
P1 115-120 Watt
P2 15-25 Watt