direct infusion fragment spectrum LC-MS/MS

Hello,

Has anyone generated a fragmentation spectrum using an Agilent QQQ (e.g., G6420A) using direct infusion?  I am accustomed to making a solution of an analyte and then infusing it directly into the ESI source using a syringe infusion pump at low flow rate (10 uL/min).  I would set the Q1 to the analyte exact mass and the Q3 to scan and then adjust the ESI source and Q2 parameters (collision energy) as the mass spec was running to generate an fragment spectrum.

Parents
  • Achem,

    You are describing a Product ion scan method in which MS1 is set to TTI (all ions allowed to pass), CC has some CE applied, and MS2 is set to scan across a defined m/z range. This can be selected as an acquisition mode in the software QQQ UI.

    If you know the precursor ion m/z, you can do this via MassHunter Acquisition software by modifying the CE in the acquisition method and monitoring via the spectrum plot. The data files will not be saved unless you run sample through the software. 

    The syringe pump will be controlled out of the MH software.

Reply
  • Achem,

    You are describing a Product ion scan method in which MS1 is set to TTI (all ions allowed to pass), CC has some CE applied, and MS2 is set to scan across a defined m/z range. This can be selected as an acquisition mode in the software QQQ UI.

    If you know the precursor ion m/z, you can do this via MassHunter Acquisition software by modifying the CE in the acquisition method and monitoring via the spectrum plot. The data files will not be saved unless you run sample through the software. 

    The syringe pump will be controlled out of the MH software.

Children
Was this helpful?