Hi fchl235
If you think about what voltages are being optimized when you tune any of the Agilent LCMS instruments, you'll see that they're all after the ions have entered the vacuum side of the instrument. Tuning should therefore be independent of the spray chamber! So, yes you can tune in ES and then switch to spray chamber you want to use.
/Andy
Hi fchl235
If you think about what voltages are being optimized when you tune any of the Agilent LCMS instruments, you'll see that they're all after the ions have entered the vacuum side of the instrument. Tuning should therefore be independent of the spray chamber! So, yes you can tune in ES and then switch to spray chamber you want to use.