Insufficient vacuum for 6210 ToF conditioning

I have recently acquired a secondhand 6210 TOF and pumped down for the first time. The installation guide states that a vacuum level of 2e-6 mbar is required to start conditioning. My current vacuum level is 1.8e-6 (no units) and only slowly decreasing, but initiating the conditioning cycle returns 'insufficient vacuum'. I am guessing that my measurement is in Torr not mbar and must therefore wait until 1.5e-6 is reached. Agilent manuals, even with the ToF range, seem to swap between these units. Has anyone encountered the same issue?

Parents
  • Yes, I've seen this before. I would probably wait until mid  e-7 anyway. The conditioning doesn't take long and you shouldn't use it until ~3e07 anyway. 

    How long has it been at e-6? You might have a leak unless its just sat out for a while. I'd think the 6120 would be down within 10 days - 2 weeks. 


Reply
  • Yes, I've seen this before. I would probably wait until mid  e-7 anyway. The conditioning doesn't take long and you shouldn't use it until ~3e07 anyway. 

    How long has it been at e-6? You might have a leak unless its just sat out for a while. I'd think the 6120 would be down within 10 days - 2 weeks. 