Chamber Current and Capillary Current

Hi,

How are the capillary current and chamber current measured?  I see that they are measured in nanoAmps and microAmps, respectively.  Are they measured through a detector?  And if so, where is this detector located?

The instrument we use is a Triple Quadrupole 6495C.  

Thanks

Parents
  • Hi Good Day,  The capillary and Chamber voltages are applied by same Power supply but the voltages are off by 500Volts and when it comes to measurement there is no special detector to measure this , When the Charged analytes are diverted towards the capillary becasue of the Electrostatic forces ( applied by V cap and V Chamber )  most of the Ions will pass and at the same time some if the Charged ions strike the spray shield and capillary Cap ( these ions will get discharged resulting drop in the voltage which power supply senses this and try to maintain the Voltage by pumping current this is measured as Nano Amps and Micro Amps) hope this clarifies . Basically this is the indirect measurement on power supply

    Regards,

    Srini 

Reply
  • Hi Good Day,  The capillary and Chamber voltages are applied by same Power supply but the voltages are off by 500Volts and when it comes to measurement there is no special detector to measure this , When the Charged analytes are diverted towards the capillary becasue of the Electrostatic forces ( applied by V cap and V Chamber )  most of the Ions will pass and at the same time some if the Charged ions strike the spray shield and capillary Cap ( these ions will get discharged resulting drop in the voltage which power supply senses this and try to maintain the Voltage by pumping current this is measured as Nano Amps and Micro Amps) hope this clarifies . Basically this is the indirect measurement on power supply

    Regards,

    Srini 

Children
Was this helpful?