Good evening. We have liquid chromatograph Agilent 1200 with time-of-flight mass spectrometric detector 6210 (Agilent Technologies, USA). When I want to connect it, it writes "could not establish telnet connection". Can you explain why?
Find maintenance, best practice, and troubleshooting articles.
Good evening. We have liquid chromatograph Agilent 1200 with time-of-flight mass spectrometric detector 6210 (Agilent Technologies, USA). When I want to connect it, it writes "could not establish telnet connection". Can you explain why?
Hello,
So it worked before ?
I would try to run the Software Verification Tool.
You said that you reinstalled acquisition. ( I would be very carefull with that ) Maybe you need to do additional steps after this.
Like all changes you make, can be critical to the system.
The best way is to contact Agilent. ( Call them )
I assume that you can reach the system with your OpenLabs ?
This problem after some time of stoping process. I opened it yesterday and can't connect. I reinstailed aquisition. It didn't help.
I reinstalled it after this issue occurred. All necessary steps made. Everything worked fine before. Last time without a problem.
try to open CMD ( Win Key + R )
then insert
ping 192.0.0.1 <--- add the ip
if this works add the port
like 192.0.0.1:23 ( If 23 is the telnet port ) im not sure
Regards
White
I opened this program, tried to connect and i can't. Maebe smth wrong with TOD diagnostic?
TOF diagnostic?
Thanks a lot. I will try.
LC is OK. I tried. But I don't understand what is wrong with communication of MS. It was good the last time of using...