Good evening. We have liquid chromatograph Agilent 1200 with time-of-flight mass spectrometric detector 6210 (Agilent Technologies, USA). When I want to connect it, it writes "could not establish telnet connection". Can you explain why?
Good evening. We have liquid chromatograph Agilent 1200 with time-of-flight mass spectrometric detector 6210 (Agilent Technologies, USA). When I want to connect it, it writes "could not establish telnet connection". Can you explain why?
I would try to run the Software Verification Tool.
You said that you reinstalled acquisition. ( I would be very carefull with that ) Maybe you need to do additional steps after this.
Like all changes you make, can be critical to the system.
The best way is to contact Agilent. ( Call them )
I assume that you can reach the system with your OpenLabs ?
I reinstalled it after this issue occurred. All necessary steps made. Everything worked fine before. Last time without a problem.
I opened this program, tried to connect and i can't. Maebe smth wrong with TOD diagnostic?
I opened this program, tried to connect and i can't. Maebe smth wrong with TOD diagnostic?
TOF diagnostic?