G6135B SQD drifting response

Hi All,

I am trying to develop a QC main peak assay for an oligonucleotide on a G6135B but I've noticed that there is a steady downward drift of response during my sequences.  The signal decrease is quite significant for the first few injections, and does slow down somewhat, but it's quite significant and is up to 15% relative to the first standard injection, which is far more than I can accept for the specification criteria.  I'd hoped that this might be an issue that would be resolved by some "conditioning" blank or standard injections at the start of the sequence, but even after 15 standard injections I still see drift.

I'm working in negative ion ESI, in SIM mode at around m/z 1800.  Eluents include water. methanol, 200mM HFIP and 2mM of a tertiary amine.

I am going to look at adding an internal standard to try to address this, but a suitable compound will be difficult and expensive to source, so if anyone had any ideas about what might be happening here or how I can address/minimise this I'd be really grateful!!

Let me know if there's any additional information that would be helpful.

Parents Reply Children
  • Hi Mitch, Danny, Richard,

    I've been meaning to follow up on this! 

    We haven't fully got to the bottom of what was happening, but it quickly became clear that it was absolutely to do with the eluent, and not the system.  Switching to an alternative mobile phase (water, MeCN, AcOH, 5mM tertiary amine) without changing anything else led to a nice stable signal.  My guess is that evaporation of HFIP from the mobile phase during the analysis might be having an impact, but we haven't done anything to prove this or to troubleshoot HFIP or the MeOH/MeCN swap further.

    Thanks,

    Chris