Hi Simon,
When you turn on the reference mass in the software, do you see it spraying from the ref sprayer in the source?
We did find a blockage in the line between the bottle and nebuliser, which we flushed with a syringe. Doing the flush has allowed us to see the reference mass again, though the abundance is 3 orders of magnitude less than expected. We are using a fresh solution from a stock preparation made in September.
The instrument was off for a period due to an unrelated failure in the lab's UPS.
I can not see any spray currently and we are thinking that the nebuliser, may also have a partial blockage.
Tubing should be replaced, part number G1969-60115 on Agilent.com.
We currently have a similar issue with our reference nebuliser on a 6546A QTOF. The reference mass solution comes through the restriction tubing fine but there is no spray and intensity of the reference ions are ~x100 lower than observed previously. I've tried a few nebulisers and no difference seen. What I did notice, however, is that the gas flow to the reference nebuliser is incredibly low, very gentle flow felt on my face.
So, I then went into tune menu of our 6546A QTOF and tried a few things.....
Is this a tubing problem or a gas flow issue?