Tuning problem with Agilent 6100

We have recently replaced the ESI needle on our Agilent 6100 LC-MS.   All is well with the flow, but we are not able to tune our instrument.  Last week, the tune looked fine, but since replacing the needle the peaks (118, 622, 922) are not found and we get a fail for the check tune.  It says to calibrate the mass axis and the peak width and it still fails when you do that.  I am confused as to why this occurred - it could just be an unwelcome coincidence but I'm wondering what to try next...I read something in the manual about cleaning the ion optics but I don't want to do that unless I'm really certain that that is the problem.   Can anyone recommend a systematic procedure to try and narrow down the problem?  Thanks.

Parents Reply Children
  • After installing new ESI tuning solution and manually tuning, I was able to get ions.  I believe the tune was completely off after changing the needle and it was unable to autotune.  After getting it tuned today, I ran a test sample (reserpine mix) and was able to get the correct mass for that.  So knock on wood, I believe the system is working again.  Thanks for your help. 

Was this helpful?