Please help advise about the part number for G1379A degasser chamber
Please help advise about the part number for G1379A degasser chamber
It depends on what you have installed. There is an upgrade kit. The old-style chambers are no longer available. You may want to contact your local service group.
Also, if too costly maybe upgrade to a newer degasser.
5067-5387 vacuum chamber replacement kit (includes everything needed, 2 chambers, tubing and sheet metal)
OR if it has been replaced- 5067-5383 (there are 2)
G1379-67310 tubing set (the old one won't work with the new chambers)
Reminder- Don't prime the vacuum degasser, using a syringe