Hello! Someone could help me?
When I set to my pump stop when level falls below to 0.1 L, my pump still go on and my system only stop when go to 0 L.
Find maintenance, best practice, and troubleshooting articles.
Hello! Someone could help me?
When I set to my pump stop when level falls below to 0.1 L, my pump still go on and my system only stop when go to 0 L.
The correct print from my pump
The correct print from my pump
Hi yurisarmento
The system is behaving exactly as designed.
The line that has "Prevent analysis" stops the system from making more injections when a bottle reaches the value supplied. The "running out of solvent" option happens when the volume gets to zero. If you want to ensure you have some liquid in the reservoirs when the system reaches zero is to set the actual volume in the table to less than the real actual volume in your solvent reservoirs.
/Andy