1200 Agilent Prep HPLC Binary Pump flow-rate problem

I use 1200 series prep hplc which includes binary pumps. The pump B gives lower flow rate with respect to the command value. I measured the accuracy which falls to app. 50% at 5 and 10 mL/min with IPA. However, Pump A works fine. I cleaned the pump head - seals, plungers, pistons, inlet and outlet ball valves. Also cleaned the filter frit, and filter cup. Pump B specifications - P/N: G1361A, Firmware: A.06.53, S/N: DE60556274. Pump A specifications - P/N: G1361A, Firmware: A.07.01, S/N: DE60555760. I appreciate if anyone can help me to find out the cause.

Parents Reply
  • Hi  .. thanks for your reply. I did replace the seals and O-ring, but that didn't make any change. The problem still continues. I assume that plungers and pistons could be the source of problems. I checked the inlet and outlet valves which looks pretty good and tested with flow of solvent using syringe. I appreciate your detailed suggestions. I will try accordingly and will let you know.