Hi
I have a 1260 Infinity SFC and the booster pump will not increase in pressure.
I have a new CO2 tank on. Looks like the booster pump is not working properly. Any recommendations to fix this problem?
Thanks!
Hi
I have a 1260 Infinity SFC and the booster pump will not increase in pressure.
I have a new CO2 tank on. Looks like the booster pump is not working properly. Any recommendations to fix this problem?
Thanks!
Hello - !
Check for frost on any of the connections which would indicate a leak. If no leaks it could be the booster pump seal is worn out. when was it changed last ?
-Howard
Hi All
It looks like the booster pump is not turning at all. The chilling plate also does not seem to be working. I'm leaning towards a blown fuse or connection to the control board?
The pump seal was changed two months ago.
Hi All,
It´s not possible replace parts individually. You should order New Booster pump drawer ( parts list with pictures for the SFC control unit - G4301A) Also I recommend you the article SFC booster pump does not turn on ( SFC booster pump does not turn on ). I replaced New booster and the pressure is turn on.
Hi All,
It´s not possible replace parts individually. You should order New Booster pump drawer ( parts list with pictures for the SFC control unit - G4301A) Also I recommend you the article SFC booster pump does not turn on ( SFC booster pump does not turn on ). I replaced New booster and the pressure is turn on.