I have a 1260 Infinity II and my needle rinse line continuously flows which is causing my needle rinse solution to run out before my sequence is completed. Is there any way this can be changed through Open Lab?
I have a 1260 Infinity II and my needle rinse line continuously flows which is causing my needle rinse solution to run out before my sequence is completed. Is there any way this can be changed through Open Lab?
Hi BJohnson
Sounds like the peristaltic pump that delivers the needle wash has a fault and needs to be replaced, assuming that you have a vial sampler or multisampler with single wash.
https://www.agilent.com/store/productDetail.jsp?catalogId=5065-4445
/Andy
__PRESENT
Yes! That is correct it has a multisampler and I see where it is continuously pumping. Thanks, I will order a new one! Is this something that can be easily changed out or would I need an Agilent Tech to be sent out?
HI BJohnson
You should be able to replace it yourself if you're happy to do that.
Have a look at page 269 in the user manual: https://www.agilent.com/cs/library/usermanuals/public/G7167-G5668-G7137-Multisampler-UseMa-en-SD-29000238.pdf
/Andy
__PRESENT