Currently i am giving blank run with injection volum 100ul and thinking that it willl clean injector
Is it appropriate way to do?
Please suggest
Currently i am giving blank run with injection volum 100ul and thinking that it willl clean injector
Is it appropriate way to do?
Please suggest
Hello Bhaskar ,
I have moved this to the Liquid Chromatography forum for better visibility.
Hey, in the driver you can right click and do a manual needle wash (to wash the outside of the needle) outside of the method with certain samplers. What module number do you have GXXXX? Flushing the flow path would need to be done by manually changing the composition (to for example a higher organic percentage) to clean the injector.