I have changed the VWD lamp, but I dont see how can I put the VWD lamp time counter on 0. I am trying from lab advisor. How can I put the VWD lamp time counter on 0?
I have changed the VWD lamp, but I dont see how can I put the VWD lamp time counter on 0. I am trying from lab advisor. How can I put the VWD lamp time counter on 0?
Garikoitz : if you install the UV lamp with RFID tag, you cannot reset the hours on the lamp. If you install the UV lamp without RFID tag, please follow the instruction from ISH. Also, did you turn off the VWD before replacing the UV lamp?
Sincerely,
Linh D.
Garikoitz : if you install the UV lamp with RFID tag, you cannot reset the hours on the lamp. If you install the UV lamp without RFID tag, please follow the instruction from ISH. Also, did you turn off the VWD before replacing the UV lamp?
Sincerely,
Linh D.