I am facing issue with G1170A valve drive - going to resident mode. i updated firmware to latest- C.07.30[001]. please suggest
Find maintenance, best practice, and troubleshooting articles.
I am facing issue with G1170A valve drive - going to resident mode. i updated firmware to latest- C.07.30[001]. please suggest
The G1170A needs a host module with either a B firmware at the same level B.3x or a D firmware at the same level D.3x. Make sure you have this module turned on with one of these other host modules. If there is no other module turned on or present, it will start in resident mode because it did not find a host.
The G1170A needs a host module with either a B firmware at the same level B.3x or a D firmware at the same level D.3x. Make sure you have this module turned on with one of these other host modules. If there is no other module turned on or present, it will start in resident mode because it did not find a host.