Hi!
I want to change the HPLC setup regarding the sample taking part, now I use sample vials, but I want to switch to 96-well plates. Has anyone done the same? Do you know who I should contact to change the HPLC setup? Do you have any suggestions or recommendations?

THANKS!

Parents Reply
  • Hey, based on the description of your container in the previous message, I think you do not have the 96DeepNunc, but rather the 96 Nunc. The dimensions could also be reviewed via the instrument configuration, then clicking on the multisampler configuration, and then define sample containers. The 96Nunc and 96DeepNunc seem to have different plate heights. 14.3 and 31.5 mm respectively. When the correct tray is chosen, the sample does everything then automatically as long as the containers have enough volume.