Hello, our Agilent 1100 pump just came back from repair. We installed and connected to Chemstation and the pump sounds correct, but solvent is flowing back and forth instead of forward. How do we start troubleshooting this?
Hello, our Agilent 1100 pump just came back from repair. We installed and connected to Chemstation and the pump sounds correct, but solvent is flowing back and forth instead of forward. How do we start troubleshooting this?
Do you have a degasser? I would try bypassing the degasser and then disconnecting at every connection to understand up to what point it reaches. Another option is if you have a syringe and an adapter, you could suck the air out of the solvent lines. You wanna use IPA for that, it is great to remove air
When the air is sucked out, sometimes it is difficult because it faces resistance. Shall the flow be on while doing this maneuver?
Yes
Yes