Hola! Como puedo saber si el inyector toma el volumen que esta seteado?
Hola! Como puedo saber si el inyector toma el volumen que esta seteado?
Translation:
Hi, how can I know if the injector draws the volume I set?
Translation:
Hi, how can I know if the injector draws the volume I set?
A couple of options that may work are:
You can weigh a capped vial containing water and then do an injection. You should lose about 1 mg/ul after using the vial for an injection.
You can use a micropipette and add maybe 20 ul of solution and then after an injection use a micropipette to withdrawal the remaining volume and estimate the volume of injection removed from the vial.