G1316C. If I try to switch from bypass to my column, it appears to have switched if I view the Column compartment in the Method Editor, but in the Instrument actuals window, it does not change. Is there a solution other than replacing the $4K valve.
Find maintenance, best practice, and troubleshooting articles.
G1316C. If I try to switch from bypass to my column, it appears to have switched if I view the Column compartment in the Method Editor, but in the Instrument actuals window, it does not change. Is there a solution other than replacing the $4K valve.
Hi, valve if not switching (where we get error for valve failed) issue is generally with column switching valve or in some cases might be the main board
Hi, valve if not switching (where we get error for valve failed) issue is generally with column switching valve or in some cases might be the main board
Hello, thanks for the response. The repair has been made. There were three possibilities. The problem was actually the actuator valve. The main board and the valve head were fine.