Bad pump that passes diagnostic tests?

Hey guys,

Our gradient test has been failing, got a new MCGV, replaced inlet and outlet filters, rotor seals, frits, switched degassers, switched lamp, cleaned flow cell, and still get a lot of drift...particularly when more than one channel is used. Pump pressure and leak tests Pass in LabAdvisor.  See attached pic.  All channels are in the same mobile phase, with varying ratios (25 to 75%) in 30 increments, A with B, C with D. Anyone ever seen this, and how to remedy?  

Parents
  • In order to check your new MCGV valve is functioning correctly, I would also suggest running this quick t-shooting test on Agilent quaternary pumps, the bubble test.

     

    Lift the bottle filter above the line of solvent in each bottle currently in use in the LC bottle tray and pump for 30 secs to introduce a bubble into each solvent line. Then place the bottle filter back into the solvent in each bottle and manually pump each solvent line separately in turn via Chemstation pump control (so 100% A, then 100% B, then 100% C, then 100% D).  While pumping from one channel only (e.g 100% A) look for movement of bubbles in the other three lines. If there is significant movement of the bubbles in the other three lines this shows you have a leak across your solvent selection valve, which would mean that your LC gradient is not being produced correctly

Reply
  • In order to check your new MCGV valve is functioning correctly, I would also suggest running this quick t-shooting test on Agilent quaternary pumps, the bubble test.

     

    Lift the bottle filter above the line of solvent in each bottle currently in use in the LC bottle tray and pump for 30 secs to introduce a bubble into each solvent line. Then place the bottle filter back into the solvent in each bottle and manually pump each solvent line separately in turn via Chemstation pump control (so 100% A, then 100% B, then 100% C, then 100% D).  While pumping from one channel only (e.g 100% A) look for movement of bubbles in the other three lines. If there is significant movement of the bubbles in the other three lines this shows you have a leak across your solvent selection valve, which would mean that your LC gradient is not being produced correctly

Children
No Data
Was this helpful?