Retention time shift on 1290 HPLC in both directions

Hello,

I would like to seek advice regarding our problem with retention time shift we get for our standards on our 1290 Infinity system. Retention time shift is in both directions to right and left and is around 0.3 to 1 % of total retention within 20 injections, which we consider as a problem as normally we observe very minimal retention shift of around 0.03 % within 400 injections. At fist we observed leaks from different ports on both switching valves in autosampler and in column compartment. All fittings were tightened and no further leaks were observed, also instrument passed all leak tests.

Do you have any ideas what could be the reason for this shift?

We know that the column is not an issue as we don't observe any retention time shift when we use column on a different system.

Best Regards

Parents Reply
  • Hello  

    Just checking if you still require assistance with this.
    If yes, the issue could be a combination of leaks and/or blockages from the valves in the autosampler and column compartment.

    If you are running Reverse phased with buffers, it may be worthwhile to replace the column with a union and flush the system with warm LC grade water and then with IPA (switch between the different flow paths using the 2 valves). If this does not work and you have spare rotor seals for the 2 valves (will be 2 different parts), you can try changing them and then perform system pressure tests to make sure they are leak tight.