Hi All, 

My lab just bought a 1260 and it was configured with Empower 3. I'm totally new with Agilent systems. I tried to create an instrument method to run caffeine. The flow rate is 1ml/min, column temp is 40C, the detection is at 273nm. There are two lamp options, one is UV lamp and the other is Vis lamp. Which one should I choose? There are also some options that are not available with Waters HPLC system, such as reference wavelength, split, and step. What numbers should I put for those options?  I made a prototype instrument method and run caffeine standard with it. I can see the peak showing in chromatogram. When I reviewed it in Empower3, the Peaks tab was missing. I was not able to label the integrated peak, and I wasn't able to see peak area, amount, tailing etc. Is this because my instrument method setup was wrong or some empower software issue?  The spectrum index tab was also missing. I think it might have something to do with Agilent DAD and Waters PDA detector. Can someone enlighten me and guide me through these problems(how to build a proper instrument method, how to get peaks tab and spectrum index show in review)? Thank you 

Nill