Hello, I have a G1321B that showed errors during the OQPV. The FSE changed the flash lamp Iand the Intensity test is OK arround 10000 counts buen the and Raman ASTM Signal/Noise fail. I want to know if there are another test to do.
Find maintenance, best practice, and troubleshooting articles.
Hello, I have a G1321B that showed errors during the OQPV. The FSE changed the flash lamp Iand the Intensity test is OK arround 10000 counts buen the and Raman ASTM Signal/Noise fail. I want to know if there are another test to do.
The intensity test looks okay for an older G1321B, still the result for the S/N test is poor. You can try using a different source of water (fresh/best quality possible) and flush the system thoroughly before running the test. It could also be a contaminated degasser, releasing algae into the system, which then show fluorescence.
If the result is still low, I would suggest to call a technician to inspect the optics (e.g. EX condenser, mirror...). Very likely the quartz window of the lamp housing needs to be replaced.