Baja sensibilidad en FLD

Buenos dias,

Tengo un detector FLD en el HPLC y veo que prácticamente no tengo sensibilidad. Conecto el lab advisor y en flash lamp life-time es de 0%. Sin embargo, el test de intensidad de la lampara da OK. A que puede deberse? como puedo solucionarlo?

Parents
  • Most important: Cleanliness is vital for an FLD. Maybe your background is too high and therefore the sensitivity dropped. Could be poor quality of water, algae growth etc.

    Also check the model number and firmware revision. On the G7121A/B older firmware had difficulties with the life-time calculation, so it stayed at 0%.

Reply
  • Most important: Cleanliness is vital for an FLD. Maybe your background is too high and therefore the sensitivity dropped. Could be poor quality of water, algae growth etc.

    Also check the model number and firmware revision. On the G7121A/B older firmware had difficulties with the life-time calculation, so it stayed at 0%.