Agilent 1290 Infinity G4220A Leak in Primary Pump Head

Hello, I am troubleshooting a leak in the primary pump head of an Agilent 1290 G4220A binary pump. I am visibly observing that when the pump is operating, mobile phase is leaking into the seal wash lines. Also, by looking at the detector signals, I can see that the seal wash liquid is leaking into the mobile phase. The Pump Leak Rate Test also fails -- the secondary piston test passes, but the primary piston can't come up to pressure, so the test aborts at that point.

I have replaced the Wash Seal (part # 0905-1175) with a brand new seal, but the problem is still persisting.

Would it be possible to get advice on what parts we should focus on replacing next? Would our symptoms mean the entire seal holder has failed (G4220-60616), or could there be other parts like the piston seal (0905-1719) or other consumable parts that could also cause the leaks we are observing between the mobile phase and seal wash fluids?

Parents
  • Hey Stacey, how does the pressure look during the primary leak rate test? Does it stay at zero (this would indicate a problem likely with the inlet valve) or it goes up and cannot reach 800 bar (this would indicate some leak inside the the chamber because of the piston or seals, or less likely a problem with the outlet valve)? I would recommend to replace the piston seal (0905-1719). Also, make sure the piston is not scratched. Also, make sure you have properly assembled the pump head. Not sure what you have, normally this will be written on the pump head for easy maintenance or longlife technology and blank for classic. Keep in mind that the older classic pumpheads require an alignment tool. Agilent 1290 Infinity Pump Head Maintenance Technical Note

    It is quite unlikely that the seal wash leaks into the mobile phase because this is behind the chamber. 

    Evaluation of pump leak rate test, see page 128: Agilent 1290 Infinity Binary Pump

  • Hello kbusch,

    Thanks for your quick response and the suggestions! I'm not exactly sure which type of pump head we have (there doesn't seem to be any labeling on the pump), but all the parts look like the ones shown on pg. 203-205 of the second manual that you linked, which seems to be the longlife? For example, our seal wash holder has the tubing inlet/outlet in the V shape, rather than parallel.

    For the leak rate test -- the pressure goes up a little bit (a few bar) but mostly stays at zero. Just to report all of what we've done:

    - The problem was originally on the Channel A primary pump head, so we worked on Channel A following pg. 160-178 in the G4220A manual. Disassembled both the primary and secondary pump heads, cleaned the interior parts, replaced both primary/secondary wash seals (but not the piston seals), and reinstalled everything to proper torque using a torque wrench. Pistons were gently cleaned with IPA and 4000 grit paper (looked smooth and unscratched), we lubricated all parts with isopropanol when reassembling, and the piston felt snug with the seal wash holder. The secondary head that we worked on is mostly OK after reassembling (there's a slight leak, ~ 3 uL/min, but not like the primary head).

    - Since we're mostly running isocratic operations (e.g. SEC), on Channel A only, we swapped the entire primary pump head for Channels A and B so that we can still run in between troubleshooting. We confirmed the problem moved to B. Here's the leak rate test:

    I've just now tried swapping the inlet valve assembly between A and B (the part with the gold seal, and also swapped the tubing coming from the degasser), but the problem is still showing up on B. Here's the leak test after swapping inlet valves:

    For the seal wash -- before attempting any of the maintenance, there was a time we were equilibrating the system with DI water and had 20% ethanol on the seal wash rinse. The seal wash was set to periodic, and we could see a peak on multiple detectors (UV, light scattering, RI, etc.) every time the seal wash pump turned on. When the seal wash was switched to DI water to match the mobile phase, the issue went away. So, I wasn't sure if the fluids were somehow mixing or if some other issue could cause that.

    Sorry for the long post, but hope that might be helpful for any suggestions on the next part to troubleshoot? 

Reply
  • Hello kbusch,

    Thanks for your quick response and the suggestions! I'm not exactly sure which type of pump head we have (there doesn't seem to be any labeling on the pump), but all the parts look like the ones shown on pg. 203-205 of the second manual that you linked, which seems to be the longlife? For example, our seal wash holder has the tubing inlet/outlet in the V shape, rather than parallel.

    For the leak rate test -- the pressure goes up a little bit (a few bar) but mostly stays at zero. Just to report all of what we've done:

    - The problem was originally on the Channel A primary pump head, so we worked on Channel A following pg. 160-178 in the G4220A manual. Disassembled both the primary and secondary pump heads, cleaned the interior parts, replaced both primary/secondary wash seals (but not the piston seals), and reinstalled everything to proper torque using a torque wrench. Pistons were gently cleaned with IPA and 4000 grit paper (looked smooth and unscratched), we lubricated all parts with isopropanol when reassembling, and the piston felt snug with the seal wash holder. The secondary head that we worked on is mostly OK after reassembling (there's a slight leak, ~ 3 uL/min, but not like the primary head).

    - Since we're mostly running isocratic operations (e.g. SEC), on Channel A only, we swapped the entire primary pump head for Channels A and B so that we can still run in between troubleshooting. We confirmed the problem moved to B. Here's the leak rate test:

    I've just now tried swapping the inlet valve assembly between A and B (the part with the gold seal, and also swapped the tubing coming from the degasser), but the problem is still showing up on B. Here's the leak test after swapping inlet valves:

    For the seal wash -- before attempting any of the maintenance, there was a time we were equilibrating the system with DI water and had 20% ethanol on the seal wash rinse. The seal wash was set to periodic, and we could see a peak on multiple detectors (UV, light scattering, RI, etc.) every time the seal wash pump turned on. When the seal wash was switched to DI water to match the mobile phase, the issue went away. So, I wasn't sure if the fluids were somehow mixing or if some other issue could cause that.

    Sorry for the long post, but hope that might be helpful for any suggestions on the next part to troubleshoot? 

Children
Was this helpful?