In our Infinity 1260 system, the robotic arm getting struck during sampling, especially in last few lines. The issue started on its own and the attached video shows the issue.
In our Infinity 1260 system, the robotic arm getting struck during sampling, especially in last few lines. The issue started on its own and the attached video shows the issue.
Any idea, how did this happened and how to prevent this kind of problem in future ?
Glad to hear your problem has been solved. Good luck
Actually It is not a common event and it is pretty rare. So the problem was just due to a rare mechanical failure
Do PM every six month and pass all test through Lab Advisor to avoid it in future..
Probably happened due to too much slack / low belt tension, that way one of the fingers gears slipped on the belt by a teeth..
So this is caused by too low of a belt tension and some event where the finger is blocked from moving, thereby slipping.
To sum up the case and provide some troubleshooting options:
Root cause:
Troubleshooting: