Robotic arm getting struck

In our Infinity 1260 system, the robotic arm getting struck during sampling, especially in last few lines. The issue started on its own and the attached video shows the issue.

Parents
  • To sum up the case and provide some troubleshooting options:

    Root cause:

    • For proper vial handling, all three fingers should be oriented concentrically.
    • In rare cases, one of the fingers might get out of position. It can especially happen when the finger gets hindered by a physical obstacle, forcing the gear on which the finger is seated to slip on the toothed belt.

    Troubleshooting:

    • Run the Gripper Verification Lab Advisor tool to verify the problem.
    • Remove the gripper using the Maintenance Positions in Lab Advisor and observe whether all three fingers are concentric.
    • If one of the fingers appears out of position (see photos above), try to force it back with your fingers.
    • If the gripper can not be recovered, replace the component (G1313-60010).
    • Rerun the Gripper Verification to verify the proper functioning of the gripper arm.
    • If the problem persists or the arm appears out of alignment, request an on-site repair from Agilent.
Reply
  • To sum up the case and provide some troubleshooting options:

    Root cause:

    • For proper vial handling, all three fingers should be oriented concentrically.
    • In rare cases, one of the fingers might get out of position. It can especially happen when the finger gets hindered by a physical obstacle, forcing the gear on which the finger is seated to slip on the toothed belt.

    Troubleshooting:

    • Run the Gripper Verification Lab Advisor tool to verify the problem.
    • Remove the gripper using the Maintenance Positions in Lab Advisor and observe whether all three fingers are concentric.
    • If one of the fingers appears out of position (see photos above), try to force it back with your fingers.
    • If the gripper can not be recovered, replace the component (G1313-60010).
    • Rerun the Gripper Verification to verify the proper functioning of the gripper arm.
    • If the problem persists or the arm appears out of alignment, request an on-site repair from Agilent.