yellow seals normal phase / black seals reverse phase

Hello!

I have a question

I have a 1290 binary pump. The default in the pump is yellow seals for the normal phase. But the mobile phase is acetonitrile/water or methanol/water for the reverse phase chromatography. Should I change the seals from yellow to black?

regards, Azrael

Parents
  • Hello Azreal,

     

    The yellow seals that come by default in the 1290 pumps are not the same normal phase seals that were used in the past.  These are the seals used for reverse phase chromatography.  From what I recall the part number is different for the 1290 seals versus the old normal phase seals used in the 1100/1200 series systems.

     

    Hope this helps.

     

    Mark  

Reply
  • Hello Azreal,

     

    The yellow seals that come by default in the 1290 pumps are not the same normal phase seals that were used in the past.  These are the seals used for reverse phase chromatography.  From what I recall the part number is different for the 1290 seals versus the old normal phase seals used in the 1100/1200 series systems.

     

    Hope this helps.

     

    Mark  