Phosphate buffer in HPLC

Goodmorning,

since few months I am using phosphate buffer for my analysis (method from Agilent Application Solution "Analysis of water-soluble vitamins from multivitamin tablets"), I've never used it before. I have an Agilent 1260 HPLC binary pump and the column is Agilent poroshell EC-C18 2.7um, mobile phases are phosphate buffer pH 7 (channel A) and acetonitrile (channel B). I am looking for a washing protocol to prevent salt precipitation.

This is what I did untill now, but I don't know where I did a mistake: the column pressure is still too high (more than 400bar with 99% H2o and 1% ACN instead of 270 bar, I've already try to restore the column with different washes,..), I have to change it (but is quite new, I've bought it last august..)

-never stop analysis letting buffer in the system (es. overnight or during the weekend)

-after analysis wash the entire system with colum with Water (HPLC grade) 90% + ACN 10% for 20 min, then return to 100% ACN

Maybe 20 min is not enough?

 

Thank you!

 

 

25/03/19

Hi guys!

i'm still here hope you can help me. i really don't know what I'm doing wrong.

I have a new column (see previous part) and after more or less 40 runs it shows higher pressure: pressure rises from 300 bar (70%acn 0.5 ml/min) to 450 bar same conditions.

Here what I did (I've used your suggestions)

- prepare daily fresh phosphate buffer [for 0.5 ml of buffer, 0.6634g KH2PO4 (mw 136.08g/mol) + 1.328g K2HPO4 (mw 174.2 g/mol)], magnetic stirrer for 15 min, vacuum filtration 0.22um.

- sample filtration with 0.22um

- for cleaning the colum: I've created a method with these conditions (ACN & HPLC water)

     0 --> 60 min 15% ACN 0.1 ml/min

     0--> 7 hours slow gradient to 100% ACN with 30 min step (each step 5% ACN more)

     other 60 min at 100% acn 0.2 ml/min

at the end the pressure was ok (less than 300 bar) but when I check the initial condition (70% acn 0.5 ml/min) the pressure became higher

 

Where is the mistake??

What should I do now? wash with isopropanol 100%? change flow? the column is too new......

THAN YOU ALL!

 

Messaggio modificato da ANTONELLA GALEONE

Parents Reply Children
  • Thanks for the pics! So I have an active one (now I know it Relaxed)

    And yes, I have no problems with other columns!!

     

    Just THANK YOU to all of you for these precious suggestions! Nice to know that I’m not alone (My company just say “use it”, but it’s not so easy, I’m not a chemist and the instrument it’s quite complicate for me..)

     

    The last question is this:

    I’ve asked a price quotation both for a column guard and an in-line filter, the price is very different ;-): in your opinion what’s the best choice in case of phosphate buffer?

     

    Antonella Galeone

    Quality Control Lab Assistant

  • I would rather prefer a guard column over an in-line filter, because you already have a mechanical filtering ptfe frit (catalog number 01018-22707) to filter everything that comes from a pump and you filter your sample with a syringe filter. But you cannot filter chemical clogging with an in-line filter (it filters purely mechanical), you can do it only with a portion of stationary phase in a guard. A guard column also has a frit by design and acts like an in-line filter with a non-replaceable frit.

Was this helpful?