PCI tune ok, but NCI tune not completing

  1. Recently we have found NCI tune not completing.  PCI tune is good as attached below, EM showing 1959 which is workable.

2. See the NCI tune. EM voltage showing high during nci tune and showing error. Is it EM issue or anything? We were having PCI tune mass issue previously. We cleaned ion source, checked filament is ok. After maintenance,  this is the situation. What should I check now?

Parents
  • NCI requires that the ion source is only 150° C. Load the default NCI tune, let the source get cool enough and wait at least half an hour for thermal stabilization and then try again. 

    What CI reagent gas is being used?  What is the gas setting in the NCI tune? Should be 40....

    Do you have an EI tune to show? Troubleshooting instrument issues in CI is difficult.  The PCI tune ion peak shapes are not symmetrical. 

Reply
  • NCI requires that the ion source is only 150° C. Load the default NCI tune, let the source get cool enough and wait at least half an hour for thermal stabilization and then try again. 

    What CI reagent gas is being used?  What is the gas setting in the NCI tune? Should be 40....

    Do you have an EI tune to show? Troubleshooting instrument issues in CI is difficult.  The PCI tune ion peak shapes are not symmetrical. 