Hello!! Is the MS40-S Vacuum pump compatible with GC/MS 7010 triple quadrupole?? Thanks!!!!!
Find maintenance, best practice, and troubleshooting articles.
Hello!! Is the MS40-S Vacuum pump compatible with GC/MS 7010 triple quadrupole?? Thanks!!!!!
The only rough pumps supported by Agilent that should be plugged into the QQQ systems are the Agilent IDP-10 and the Edwards RV-5. Those are the only tested and supported pumps and the only ones used for Agilent service calls. If the MS40-S has the same or better pumping capacity it will work connected by the vacuum hose, of course, and the safest would be to plug the power into a wall outlet to avoid any potential current draw issues on the QQQ mainframe.
The only rough pumps supported by Agilent that should be plugged into the QQQ systems are the Agilent IDP-10 and the Edwards RV-5. Those are the only tested and supported pumps and the only ones used for Agilent service calls. If the MS40-S has the same or better pumping capacity it will work connected by the vacuum hose, of course, and the safest would be to plug the power into a wall outlet to avoid any potential current draw issues on the QQQ mainframe.