Acquisition Method Error - Open Lab CDS

Hi all, 

I am using using Open Lab CDS and am having an issue with setting up the SIM acquisition method. I have selected 74m/z as the unique ion to this compound group. For the fatty acids which do not include 74m/z as part of their spectra, I would like to select their characteristic ion to monitor. I am able to do this for the majority of the compound list, but for some when selecting plot ion I receive the following error "The number of signals displayed in the real time plot is 13 and is limited to 12." When I deselect this ion, the error is resolved. Any ideas on how to resolve this issue? 

Thanks!

Parents
  • Yes. Don't select that many ions to plot in real-time plot as shown on the screen during the run. The ions will all be saved in the data file for you to see there.  Most only plot a very few ions.   

    You have many SIM groups.  If your retention time changes, you'll have to fix the times for all of them.  You may want to consider condensing them a bit?  Like 43 and 43.7 together with 6 ions, and maybe 51 and 51.9 together with 6 ions.   Like that...

    It's important to balance the number of ions and the dwell time per ion to allow enough data points to make useful peaks.  If you have few ions, like the three or four shown, then you can use longer dwell times than some do.  You want to have at least three to four SIM ion cycles per second so that you get at least ten to twelve or more data points for each chromatographic peak. Mathemeticians say it takes that many point per peak to make one that can be integrated properly and used for quantitation.

Reply
  • Yes. Don't select that many ions to plot in real-time plot as shown on the screen during the run. The ions will all be saved in the data file for you to see there.  Most only plot a very few ions.   

    You have many SIM groups.  If your retention time changes, you'll have to fix the times for all of them.  You may want to consider condensing them a bit?  Like 43 and 43.7 together with 6 ions, and maybe 51 and 51.9 together with 6 ions.   Like that...

    It's important to balance the number of ions and the dwell time per ion to allow enough data points to make useful peaks.  If you have few ions, like the three or four shown, then you can use longer dwell times than some do.  You want to have at least three to four SIM ion cycles per second so that you get at least ten to twelve or more data points for each chromatographic peak. Mathemeticians say it takes that many point per peak to make one that can be integrated properly and used for quantitation.