Hello i asked earlier about HED Fault i encounterd
The Autotune gives error
Filament turns off
I check for air leak ... the door closes tightly immediately when the vaccum pump turns on
Hi Vac is N/C
And this is the error
Hello i asked earlier about HED Fault i encounterd
The Autotune gives error
Filament turns off
I check for air leak ... the door closes tightly immediately when the vaccum pump turns on
Hi Vac is N/C
And this is the error
Hi @Mg7, there are the possible causes:
1)make sure the solvent delay is long enough
2) Vacuum is not good enough like the vacuum manifold pressure is too high
3) HED high voltage lead is shorted to ground. please check the HED high voltage lead. Make sure it is connected and not shorted to a grounded surface.
4) the detector is contaminated, Cleaning the HED - 5973A/N (agilent.com)
5) MSD electronics, it could be a problem with hed, hed power suplly or main board
I appreciate your feedback,
Thanks
Hi @Mg7, there are the possible causes:
1)make sure the solvent delay is long enough
2) Vacuum is not good enough like the vacuum manifold pressure is too high
3) HED high voltage lead is shorted to ground. please check the HED high voltage lead. Make sure it is connected and not shorted to a grounded surface.
4) the detector is contaminated, Cleaning the HED - 5973A/N (agilent.com)
5) MSD electronics, it could be a problem with hed, hed power suplly or main board
I appreciate your feedback,
Thanks