I have a problem with the transfer of set data in the MassHunter GC / MS Acquisition program to Agilent 8890 GC.
The problem occurs with most, but not all (!!), loaded methods.
For example I load the problematic method with set 'Oven Temp On' on 70 C but the Oven Temperature do not decrease from 80 C to 70 C (and Agilent 8890 GC do not reacting on the change too).
It should be done automatically by loading method or press 'ok' button then I change the data but it's not.
![](/resized-image/__size/1280x960/__key/communityserver-discussions-components-files/147/2022_2D00_10_2D00_14-15_5F00_21_5F00_39_2D00_Photos.png)
But then I changed it manually on the Agilent 8890 GC screen the temperature goes down and it works fine.
![](/resized-image/__size/1280x960/__key/communityserver-discussions-components-files/147/agilent2.jpg)
The same issue is with the rest data like 'Flow' or 'Pressure' and the rest when I load a problematic methods.
![](/resized-image/__size/1280x960/__key/communityserver-discussions-components-files/147/2022_2D00_10_2D00_14-15_5F00_30_5F00_00_2D00_Photos.jpg)
Can you advise what is wrong?
Best Regards