Front Inlet constant leak

Hi,

I recently cleaned a 7820 GC inlet and replaced all the replaceable parts. After starting the GC I realized that I have inlet leak. When I start the leak check total flow stays on ~3.4 while column flow is ~1.3. I checked all inlet parts, reopened and closed them, but still the flows remain the same. It doesn't even increase, let alone decrease.

Any help would be really appreciated!

Ehsan

Parents Reply
  • I removed the column and closed the lower end of the inlet and I saw a 1ml reduction in the leak. After that I removed the liner and inserted a septum instead (image) (I also put a new septum in the injection port and checked the leak using soup). Literally, I removed inlet from the system, but the leak still was there. As of now, the total flow is 2.4 ml and the column flow is 1.4 ml. I think the gas controller may have leaks.

Children