Hello,
with our new GC/MSD 5977B always comes the message:
Front inlet Pre-run flow test failed, Run Contiuing
or
Front-inlet: Pressure Pre Shutdown warning
We have just recently replaced the gas filter. Can this be the cause?Maybe a leak?
Find maintenance, best practice, and troubleshooting articles.
Hello,
with our new GC/MSD 5977B always comes the message:
Front inlet Pre-run flow test failed, Run Contiuing
or
Front-inlet: Pressure Pre Shutdown warning
We have just recently replaced the gas filter. Can this be the cause?Maybe a leak?
Hi
May be there is leak..
I assume that you vent the system to change gas filter so tight septa nut and column nut again little bit..Generally, it leaks at there...
Hi
May be there is leak..
I assume that you vent the system to change gas filter so tight septa nut and column nut again little bit..Generally, it leaks at there...
Thank you! We will test your advice :-)