I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
Hi
Did you check PFTBA level.
Did you check column length.
May be pftba is low or column length coming out from transferline in MS is more.
How old is Electron multiplier horn..
dear ISH
checking column length was 1 mm after tip.
checking PFTBA level. was ok
Electron multiplier horn. is 3 years old and give emv 1600
Okay..
Can you share last tuning report..
Okay..
Can you share last tuning report..
You are sure that no one did any dope frequency test..
dope means what?
Means Quad frequency adjustment.
Check again all wires are connected properly otherwise it looks electronics problem to check onsite..
i sent a pic to tune file
i already checked all wire and all ok
dipping the Quad did not done
this problem happen after electricity cut off from the building while the instrument was running