Using an AUX EPC for helium pressure control at the inlet of a Markes Thermal Desorption unit

Hello everybody,

Our GC-MS has two setups:

- classic: GC injector - column - MSD,

- less classic: thermal desorption (TD) unit - transfer line until GC oven - ultimate union - column - MSD.

The less classic setup operates at constant helium pressure controlled though a manual valve and a gauge installed before the TD unit and after the gas filters. There is the need to swap columns between setups as they are different.

We have an unused AUX EPC installed (I think it was meant for a purged union for backflushing purpose) and I was wondering if I could connect one of the outlets of that EPC to the inlet of the TD unit to start controlling the pressure though the GC-MS software. This way I would have a better control of the method pressure and, most importantly, I could leave both columns in the oven no matter what setup I'm using (I would always control a small pressure of He in the unused column).

However, I have been told that it could not be that kind of EPC. It had to be a Split/Splitless or Multimode EPC.

Can someone explain what is the difference between those EPC's?

Thank you.

Cheers.

Parents Reply
  • So it’s always best to have split on in standby to keep a positive flow of helium through the heated valve and out of the split vent for low air and water on MS. if you have a system with mass flow control of trap and split vent flows then I’d recommend 5-10 ml/min. If you don’t and only have needle valves then I would still leave split on in standby but you’ll have to manually adjust down the split flow when you’re leaving the system in standby.