Hi,
Which is the correct flow in saver gas for GC/MS? I have it at 20ml/min.
Hi,
Which is the correct flow in saver gas for GC/MS? I have it at 20ml/min.
Hi Carol,
That is perfect. The gas saver defaults to that flow and the design of the saver feature was for 20/ml/min which allows enough total flow to operate properly.
Regards
James
Hi Carol,
That is perfect. The gas saver defaults to that flow and the design of the saver feature was for 20/ml/min which allows enough total flow to operate properly.
Regards
James